!!

Welcome to our Forum, extension and content sharing platform of Electrical Test and Measurement Product Reviews Blog.

 
Registration this is Free, we accept contributors of 17 years old and above. We do not accept registration with Gmail

Active since 7 June 2012  


Copyright Notice: Entire ITTSB.EU content & images they are copyright protected. - Forum search engine disabled to Guests - No need of you using Adblock software.

Author Topic: Negative Ohms Readings? Your are victim of Temco (Temperature coefficient)  (Read 2558 times)

0 Members and 1 Guest are viewing this topic.

Offline Kiriakos GR

  • Administrator
  • Hero Member
  • *****
  • Join Date: May 2012
  • Location: Greece
  • Posts: 2202
  • Country: gr
    • ittsb.eu
  • job title: Industrial Maintenance Electrician
I came across negative Ohms readings by testing calibration of three Keysight U1272A multimeters and I stayed puzzled for several minutes.
The answer came to me by looking these meters LCD which at top right side there was the indication of temperature.

Greece in November translates to beginning of true winter, and at that day room temperature was low at 16~17 Celsius.
All test and measurement equipment's they get calibrated at 23C room temperature for a good reason.

Negative Ohms Readings in the range of mOhm this is side effects influencing   measuring circuit because the meter is not warm enough as it was at the time of calibration.

Zero Ohm value this is resistance at 23C by the law of Temco (Temperature coefficient).
When ppm of resistance decreases a regular multimeter will display negative Ohms value by design, because there is no other way to inform us that few ppm are missing from what we define as zero ohm resistance at 23C.

Basic Physic lessons at school it clearly teach us properties of metals, cold metals equals to less resistance, hot metals equals to higher resistance.
 
Temco (Temperature coefficient) at multimeter it can be more obvious when a cheap meter operates at 15C or bellow that point, and all that it is needed is you to sort your test leads ends at resistance smallest range.
Temco (Temperature coefficient) at multimeter at DC voltage this is a different story, no one can see a reading of missing micro-volts because HH multimeter does not have such high display resolution.

If you are calibration enthusiast and you are in need testing your DMM then you should wait to perform such tests in a warm day or in temperature controlled environment.  ;)
WWW.ITTSB.EU   Industrial Test Tools Scoreboard  (Product Reviews Blog) / Editor in Chief.
The content of this Web site is copyright protected

Offline Kiriakos GR

  • Administrator
  • Hero Member
  • *****
  • Join Date: May 2012
  • Location: Greece
  • Posts: 2202
  • Country: gr
    • ittsb.eu
  • job title: Industrial Maintenance Electrician
Five months later than this topic creation, and my room temperature this raising now close to 20C and just a bit above, soon it would be a good time to perform a new inspection of all three U1272A and also of my newcomer 6 1/2 precision multimeter.     

Specific 6 1/2 precision multimeter this has voltage reference with active heater and this translates that it is immune to Temperature coefficient drift.
Therefore in the upcoming weeks I will have the opportunity to explore how immune really this is … At 23 Celsius and higher than that.   :)
WWW.ITTSB.EU   Industrial Test Tools Scoreboard  (Product Reviews Blog) / Editor in Chief.
The content of this Web site is copyright protected

 

ITTSB.EU Blog

General Data Protection Regulation GDPR ITTSB.EU Home Page Reward us by a Donation - Sponsorship TsDMMViewer Data Logger for FLUKE 884xA

ITTSB Blog - Sponsors

protosnet.com - Internet solutions FLUKE benchtop DMM Repair Services

Recommended Links

Hellenic Accreditation System E.SY.D. Portal of city Volos - Greece Clean Up and Customize Facebook Firefox Backup Tool 32bit / 64 bit Winaero Tweaker = Win7 Fonts size fix