Fluke corporation in most of their DMM it does deliver this somewhat of generic advice, as Current Input Fuse Self-Test.
With the Meter powered up, plug a test lead into the
V/Ohm/diode connector (banana plug).
At testing of
400 mA input, If the fuse is good, the Meter will read less than
200 Ohm.
At testing of
10 A input, If the fuse is good, the Meter will read less than
1 Ohm.
(My own question which does not find an answer, this is if FLUKE 8845A 8846A, them are able to detect a deteriorated Fuse?)
For the owners of 8845A & 8846A, I will add reference images of what actually my DMM measures at both tests.
Luckily I did find another photograph in China of 8846A, at 10A test, but the digital camera took an incomplete photograph.
My plan is this topic to become a holder of statistics, but I cannot do that alone, we need the input from several users.
Even so I will make the first step, and at my next message I will deliver actual measurements due my own DMM.