Two days before
New year 2026, and I started running verification tests at my 8846A.

Within
2025, my significant investment this was getting
a real single phase power meter.And I totally forgot to perform a fresher power consumption AC Mains test at my 8846A.
PM9913 this is specialized AC power meter for measurements of
very low AC power. As it is demonstrated bellow, I did run two quick tests, 8846A soft power On and OFF.
The measured differences in Watts Power, these now clearly looking at much higher precision than the old test.
New test at
16C room temperature, and
12.40W this my latest and accurate measurement.
The
low cost power meter 10A range, this had measure
13.2W = 0.8 Watt higher.
At
soft Power OFF mode, the new accurate value this is
9.168W , while previously I had measure
stand by at
9.5W, = 0.33Watt Error.

I have learned allot in the year 2025, about
AC power measurements, and the reason of why the inexpensive power meter shown small error, this is the copper transformer within the 8846A.
The copper transformer elevates the
Power Factor at a point were the low cost power meter, this can measure with minimal error.
The test results they would be much more different, if 8846A this was using
a switching power supply instead.

The measurement of
Low Power Factor, this is not possible due 10$ worth power meter.
PM9913 this is also capable to display
VA and
VAR, at this test, I selected
VA.
PM9913 this is also able to perform averaging due PC software, but I do not think that
FLUKE Corporation this is interested for a complete Power consumption printed report according to IEC.

In round numbers,
12.40W =
16.18VA at
235V Input.
I did not use stabilized input voltage ACV, but I got lucky, both high and Low watt measurements they were captured at identical conditions.