Its not the first time that
FLUKE 8846A documentation, this found to be at lowest standards.
DIODE Test Mode this is another poorly documented function in English.
Over the years FLUKE Corporation in the USA, this share the duty of writing of localized marketing material at several FLUKE owned offices around the globe.
Most well made documentation, its paper work made by
FLUKE Japan !!Over such a documentation, I found also the note that
8845A/8846A they were designed to test all sort of diodes, LED and even
Zener diodes up to 10V.
Today neither
AI this is aware of the actual capabilities
of voltage and current source over the 8846A.

8846A open-circuit voltage at
5V mode = 6,140 VDC (1mA output & 10,1
M Ohm DMM input)
8846A open-circuit voltage at
10V mode = 12,825 VDC (1mA output & 10,1
M Ohm DMM input)
8846A current source
1mA = 1005,0 uA ( 0~400mA HH DMM input)
8846A current source
0.1mA = 100,68 uA ( 0~400mA HH DMM input)
Test performed at
16C environmental.
Current source
0.1mA, this is for special low power high precision diodes, that electricians will never touch.
Lets check what competitors do ... Truevolt Series by KS, three models at
5.05V Max
KS 3458A Up to
12 V (mode dependent)
Legacy KS 34401A model, which only provided
1.2V Keithley 2000, , including 3V and even a high-voltage 10V range for specialized testing.
Thanks God that FLUKE retired
8845A/8846A and made all their competitors to smile again.