Today May 10 2025, due my active communication with
Mr. Dumitrescu Application Engineer,
we both came in to a dead end.

We both verified due lab testing, that the information supplied due
Keysight R&D engineers
back to us, this is just a theory of operation.
While the DMM, the hardware,
this is not responding according to this theory of operation.
Long story in sort,
Keysight U127xA User manual documentation, this implying that
Smoothing function ... this might be assisted by the
Ahold Setting.
In Theory: Smooth setting
0007 along with
Ahold Setting
0500 =
Two adjustments and or, two points of control, these as combined... will help the DMM display, at the delivery of a more stable reading.
..............................................................................................
At ITTSB lab, I did run tests with U1272A.
Mr. Dumitrescu he run tests with U1271A.
We both came down in to identical conclusion.
a)
Smooth settings range
0007 up to
0050 , these offer a minor improvement.
b)
Ahold Settings range, no matter the input value, these are ignored by the hardware.

---------------------------------------------------------------------------
In conclusion, something when very wrong withing Agilent / Keysight, back in time that U127xA series this was still under design / development.
Human error of other?
Smooth this does not work as expected at the specific DMM series.
Final word of
Keysight R&D,
Consider AVG Feature: If the environment has significant noise, the AVG feature could help stabilize the readings more effectively.
AVG Feature = Min/Max DMM Function.